Endotracheal tube cuff pressure during flight

Endotracheal Tube Cuff Pressure before, during, and after Fixed-Wing Air Medical Retrieval

Prehospital Emergency Care: Posted online on December 19, 2012.

Increased endotracheal tube (ETT) cuff pressure is associated with compromised tracheal mucosal perfusion and injuries. but no published data is available for Australia on pressures in the fixed-wing air medical retrieval setting.

© Gary Wilson/ Pre-hospital Research Forum

© Gary Wilson/ Pre-hospital Research Forum

After the introduction of a cuff pressure manometer at the Royal Flying Doctor Service (RFDS) Base in Dubbo, New South Wales (NSW), Australia, the authors assessed the prevalence of increased cuff pressures before, during, and after air medical retrieval.

This was a retrospective audit in 35 ventilated patients during fixed-wing retrievals by the RFDS in NSW, Australia. Explicit chart review of ventilated patients was performed for cuff pressures and changes during medical retrievals with pressurized aircrafts. Pearson correlation was calculated to determine the relation of ascent and ETT cuff pressure change from ground to flight level.

It was found that the mean (± standard deviation) of the first ETT cuff pressure measurement on the ground was 44 ± 20 cm H2O. Prior to retrieval in 11 patients, the ETT cuff pressure was >30 cm H2O and in 11 patients >50 cm H2O. After ascent to cruising altitude, the cuff pressure was >30 cm H2O in 22 patients and >50 cm H2O in eight patients. The cuff pressure was reduced 1) in 72% of cases prior to take off and 2) in 85% of cases during flight, and 3) after landing, the cuff pressure increased in 85% of cases.

The authors believe that the high prevalence of excessive cuff pressures during air medical retrieval can be avoided by the use of cuff pressure manometers.

http://informahealthcare.com/doi/abs/10.3109/10903127.2012.744787